TROSY
The chemical shift resolution and sensitivity of NMR generally benefit from an increase in magnetic field strength. As a result, large sums of money are spent on magnets with higher and higher fields. The boost in sensitivity means that smaller and smaller quantities of sample are needed and measurements can be completed in shorter periods of time. There are particular cases however, where an increase in magnetic field can lead to a loss of sensitivity and resolution. This is the case for 15N decoupled 1H spectra of the 1H-15N spin pairs in very large 15N labelled proteins. The very long correlation times of the protein combined with the high resonance frequencies associated with high field strength lead to very short
T2 relaxation times and therefore broader lines. The broad lines account for a significant loss in resolution and sensitivity. One might then wonder why protein structural chemists spend so much money on very high field magnets. What follows is one possible answer to this question.
The two main relaxation mechanisms for the 1H and 15N in proteins are dipolar coupling and chemical shielding anisotropy. These two mechanisms are also cross correlated with one another. The cross correlation term is of different sign for each of the two peaks in a 1H-15N doublet resulting in one of the peaks of the doublet having a shorter
T2 (and broader line) than the other one. If 15N decoupling is applied, one sees a single resonance with a line width determined by the average of the two components of the doublet. This is illustrated in the figure below for a 1H-15N spin pair in a small and large molecule at high field . The same is true in the 15N-1H doublets in the 15N spectra of 15N-1H spin pairs.
At very high fields, One of the lines in the 1H-15N doublet is very sharp and the other very broad. If, in the 1H spectrum of a protein, we could eliminate all of the broad doublet components leaving only the sharp ones, we would have a high resolution 1H spectrum. Further, if we could combine such a measurement with an
HSQC, we would have a high resolution 1H-15N
HSQC at high field. The combination of these two measurements is called
transverse
relaxation
optimized
spectroscop
y (TROSY). TROSY data collection employs an
HSQC measurement with neither 1H nor 15N decoupling elements (as described in a
previous post) as well as other elements which suppress the broad lines of the doublets and retain the sharp lines. The results of this are illustrated in the figure below for small and large proteins at high field.
Clearly, it is not advantageous to use the TROSY technique on small proteins rather than the conventional
HSQC. For large proteins at high field however, there is a significant sensitivity and resolution advantage compared to a conventional
HSQC. It should be noted that the TROSY cross peaks are shifted by ½ 1JHN in both the F2 and F1 domains. The figure below shows a superposition of a conventional
HSQC (black) and a TROSY (blue) for a protein at 500 MHz.
One can clearly see the ½ 1JHN shift in the F2 and F1 domains of the TROSY compared to the
HSQC. In this case, the conventional
HSQC gives higher sensitivity than the TROSY.
Thank you to
Adam Damry of
Professor Roberto Chica’s research group at the
University of Ottawa for providing the sample of 15N labelled protein.
Source:
University of Ottawa NMR Facility Blog