Fast 90 Degree Pulse Determination
Almost all NMR measurements rely on the correct calibration of 90° pulses. This is traditionally done by collecting a series of spectra as a function of pulse duration, finding a null for the 180° or 360° pulse and calculating the 90° pulse by simple division by 2 or 4 in the case of the 180° and 360° nulls, respectively. This determination, although trivial, can be very time consuming. Wu and Otting* have presented a much faster method of determining a 90° pulse based on measuring the nutation of a magnetization vector directly. Continuous nutation is depicted in the figure below.
Here, the sample is subjected to continuous irradiation about the x axis. While being irradiated, the magnetization vector rotates in the z-y plane at a nutation frequency proportional to the pulse power. The magnetization on the -y axis is defined by a sine function. Fourier transformation of this magnetization gives an antiphase doublet centered at zero whose splitting ?
? is twice the nutation frequency. The reciprocal of the nutation frequency is the time it takes the magnetization vector to rotate one complete cycle in the z-y plane and therefore the time it takes to rotate by one quarter of a cycle (i.e. the 90° pulse duration) is defined as 1/(2 ?
?). The problem with continuous irradiation is that the sample must be irradiated at the same time magnetization is being detected. To eliminate this problem, a scheme similar to
homonuclear decoupling is used where the radiation is turned off long enough to sample a data point. This is depicted in the figure below.
Here each dwell period is divided up into a period for irradiation and a period for detection. The duty cycle for the irradiation is the fraction of time for which the sample is being irradiated. The magnetization is sampled when the power is off. As in the case for continuous irradiation, the magnetization vector still rotates in the z-y plane however, the rotation is slower as it is scaled according to the duty cycle. The duration of the 90° pulse is
d/(2 ?
?), where
d is the duty cycle for irradiation. An example of this is shown in the figure below.
The nutation spectrum was measured for HDO using a duty cycle,
d = 0.10 and a power level of
12 dB (Bruker). Since the response of the amplifiers is linear, the 90° pulses at higher power levels can be calculated. Each decrease by 6 dB cuts the duration of the 90° pulse in half. In this case the 90° pulse at 0 dB was calculated to be 10.93 µsec at 0 dB based on the measured 90° pulse of 43.71 µsec at 12 dB. This pulse agrees to within a couple of percent of that measured by the more traditional method however, the measurement took only a few seconds. You can use a program called "pulsecal" on newer Bruker spectrometers to do this in complete automation.
--
* Peter S.C. Wu and Gottfried Otting
J. Mag. Res.
176, 115 (2005).
Source:
University of Ottawa NMR Facility Blog